saoye wrote:
The only difference I see is the meterAdjustAdds=17 What does that actually do?
When I piggyback the CS provided Meter off the back of the native Proaster thermocouple, the CS meter temp is not accurate. (The Proaster shows the correct temp so I can see the differences). The amount that it is out, (and always lower), varies to some degree (no pun intended), but when it reaches the temps suitable for roasting it stabilises. The meterAdjustAdds=17 calibrates the CS Monitor Software so that it shows the (virtually) correct temperature during the roast cycle, so can be used to profile and track. The following is the description of the Meter calibration factors for the CS Monitor software: Meter calibration factors: You can modify the displayed values of the 9 meters by setting values in preferences.txt. This can be used either to precisely calibrate a meter or to allow a reading other than temperature (e.g. airflow) to be scaled to suit the graphs. Each meter has a multiplier (meter1AdjustMult, meter2AdjustMult. .......etc ) and an offset (meter1AdjustAdd, meter2AdjustAdd, .........etc). For example, if meter 2 is measuring airflow that registers 4 mA at 0% and 20 mA at 100%, you could set: meterAdjustMults=1.0 6.25 1.0 1.0 1.0 1.0 1.0 1.0 1.0 meterAdjustAdds=0.0 25.0 0.0 0.0 0.0 0.0 0.0 0.0 0.0 so that it would display as values between 0 and 100. In other words, for the 2nd meter the multiplier (6.25) extends the initial range of 16 to a new range of 100 (25 - 125); while the offset (-25.0) adjusts the zero position. Hope that is helpful Graham